Hello,
Please refer to below steps. You need a DC adjustable Power Supply, multimeter and load
If the no-load current is not zero, please follow the zer calibration firstly.
Zero calibration:
- Please hold the OK button and turn on power supply simultaneously. Then release button after 3 seconds, it will come to parameters setting.
- Please make sure load is disconnected and yellow cursor is at “Mode”option, then click “OK” button to move yellow cursor to “0A”, then short press Down button, current reading will return to zero.
- At last, short press OK button to move cursor to “Save”option to save the setting.
Charging current calibration:
- Hook up meter, power supply and load asper charging wiring diagram.
Set the power supply as above12V, then hold on the Ok button and power on. Release button OK after 3 seconds , it will come to parameter setting interface.
- At that moment, yellow cursor is at “Mode” , the third line of interface show “00000”. Please disconnect load or just make sure load current is 0.00A, then press OK button to move the cursor to “ 0A” ; Then press down button, the displayed current value will return to zero
- Press OK button to move to”+4.A”, then connect load, the load current must be above 1.00A; Then press up or down buttons to make the measured value is the same as load current.; Next, press OK button and move to ‘Save’to save the calibration setting, then power off and power on again.
Discharging current calibration.
- Hook up the power supply and load according to the discharging wiring diagram.
- Set the power supply as over 12V, then hold on OK button and power the meter simultaneously.Then release button OK after 3 seconds. It will come to parameter setting;
- Please move yellow cursor to “Mode”and short press OK button, then move the yellow cursor to “-4.A”.
In that case, please make sure the actual load current is over 1.00A. Next, please press up or down button to adjust the current value of meter until it is the same as actual load current. After that, press OK, move yellow cursor to ”Save” and save the calibration.
Bjorn SjolingDecember 30, 2021 at 11:29 amPost count: 0Hello,
We don’t sell 500v version, please check if it is this item, https://www.amazon.com/dp/B07B1SJ8W6?ref=myi_title_dp
Best regards.
greg jarvisJuly 6, 2022 at 8:06 pmPost count: 0Hello greg jarvis,
Please refer to below cablibration instruction
- Connecting power supply (20V/1A) and load. (It must be 20V/1A)
- Short connect two calibrating spots (Open the back cover, you can find two holes which
marked with “W” on board)
- Turn on the power supply.
- When the display changed from “CAL—“ into” —PASS”, that means the calibration is finished.
-